Tejoh
Login Register

System suitability in UV Spectrophotometer Methods


0
Q
Queries Qvents
2026-03-09 06:04:50

What criteria can be defined for system suitability in UV Methods for Assay, Impurities (Limit test) by UV-Vis Spectrophotometer

0
I
InfoQvents
2026-03-09 06:08:57

Pharmacopeia do not define specific system suitability tests for UV methods. However the following may be considered.

  1. A test of repeatability by measurement of response of a standard solution, for example response of standard in five replicates (with %RSD to be less than 2.0%)
  2. Resolution with a solution of toluene in hexane, criteria - ratio between the minimum and maximum absorbance at respective wavelengths of 267 and 269 nm must exceed 1.3 (USP)